-----Original Message-----
From: stoves-bounces at listserv.repp.org
[mailto:stoves-bounces at listserv.repp.org] On Behalf Of Tami Bond
Sent: 04 February 2008 02:27
To: Discussion of biomass cooking stoves
Subject: Re: [Stoves] Making testing equipment for CO and O2

Hi all,

I *think* that Nernst cells are the same principle as the electrochemical
cells used in most inexpensive CO sensors. However, some of the
electrochemical cells are corrected for interferences (measuring one gas as
another).

I can see a need for measuring O2 in the exhaust of a chimney stove, or the
outlet of a non-chimney stove. O2 is available in most combustion analyzers.
I'm much more in favor of measuring CO2 because (1) it can be directly
traced to the fuel and (2) if you have a lot of excess air/dilution, as you
do above a non-chimney stove, the O2 quickly approaches atmospheric content
and the calculation of excess air becomes very sensitive to small
differences. The calculation of excess air is fairly robust if you are
measuring 5% or 10% O2, but not very robust at all if you measure 19% O2.
That's why we like to measure CO2 directly, and then the CO/CO2 ratio is
better known. We get our CO2 sensors
(infrared) from Telaire and they do take some wiring and calibration. 
These inexpensive CO2 sensors have been available only in the last ~5 years,
and they are slowly making their way into combustion analyzers. 
O2 is still far more prevalent, but again, I think it is much less
appropriate to stove applications.

Tami

*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*-*
Assistant Professor
Arthur & Virginia Naumann Endowed Faculty Scholar Dept. of Civil &
Environmental Engineering University of Illinois, Urbana-Champaign
Newmark Civil Engineering Laboratory, MC-250   (01)217-244-5277
205 N. Mathews Ave., Urbana, IL 61801   USA    fax/217-333-6968



_______________________________________________
Stoves mailing list
Stoves at listserv.repp.org
http://listserv.repp.org/mailman/listinfo/stoves_listserv.repp.org
http://stoves.bioenergylists.org
http://info.bioenergylists.org




More information about the Stoves mailing list